IowaCowboy
Want to go back to Iowa
Premium
join:2010-10-16
Springfield, MA
Reviews:
·Comcast
·Verizon Broadban..

UPS back-up for a CPAP machine

Click for full size
UPS I am thinking of buying
I am thinking of purchasing a UPS back-up power supply to plug my CPAP machine (medical device for sleep apnea) into as I use a full face mask and worry about an interruption in the power supply causing me to suffocate. I am looking at the UPS pictured at Best Buy. I would invest in a standby generator but considering the house is a rental and I do not want to invest that kind of money into a property I don't own.

I have severe sleep apnea to the point of being life-threatening.

Here are the power specifications taken off of the CPAP manufacturer's website. I also use it with a heated humidifier that is attached to the machine and is powered by the machine.
90W power supply unit
Input range 100–240V, 50–60Hz,
Nominal for aircraft use 110V, 400Hz
Typical power consumption 70W (80VA) Maximum power consumption 110W (120VA)

UPS Specs: 865 Watts, 1500 Va. up to 164 minutes run time.


DarkLogix
Texan and Proud
Premium
join:2008-10-23
Baytown, TX
kudos:3

Well I'd consider the following
1. how long do you normally sleep? I'd bet its over 3 hours so if you lost power shortly after getting to sleep then it wouldn't last, though the beeping might wake you up.

2. APC's backup's line only adjusts so fat based on load, ie even on a tiny load you won't get over 164min

3. APC's smart ups are much more able to adjust to the load, and the Smart-ups XL UPS'es can have an add on battery to further extend run time

all that said the smart ups models cost much more, and the beeping of the Back-ups might be enough to wake you in case of power failure

I'd get the one pictured and test it before depending on it.



hitachi369
Embrace Your Rights
Premium
join:2001-10-03
Grand Rapids, MI
kudos:4
Reviews:
·AT&T U-Verse

reply to IowaCowboy
I accept no liability for anything, you should consult your medical advisor.

The UPS will beep when the power goes off, so run time likely won't matter, as for me at least the beeping wakes me up. At which point you would be woken up and can decide what action to taken.



ArthurS
Watch Those Blinking Lights
Premium
join:2000-10-28
Hamilton, ON

reply to IowaCowboy
Run time on a UPS battery is also dependent on the pressure your CPAP machine is set at. Have you explored battery operated CPAP machines? There are several options available here:
»www.cpap.com/cpap-battery-power/···tery.php

You might want to visit their forum as well, lots of helpful people there that have the same challenges as you:
»www.cpaptalk.com/CPAP-Sleep-Apnea-Forum.html



Draiman
Let me see those devil horns in the sky

join:2012-06-01
Kill Devil Hills, NC
Reviews:
·Verizon FiOS

reply to IowaCowboy
Just go buy the APC. Take it home then hook up the unit and let it charge up for 24 hours. Next hook up your machine and pull the plug on the APC. Start a stop watch and see how it works. If it works well your set. If not put it back into the box and return it to Best Buy.



cdru
Go Colts
Premium,MVM
join:2003-05-14
Fort Wayne, IN
kudos:7

reply to IowaCowboy

said by IowaCowboy:

I am thinking of purchasing a UPS back-up power supply to plug my CPAP machine (medical device for sleep apnea) into as I use a full face mask and worry about an interruption in the power supply causing me to suffocate.
...
I have severe sleep apnea to the point of being life-threatening.

Have you ever heard of someone dying from suffocation due to a power outage while on a CPAP machine? They are designed for such instances. Worst case, if the power goes out, you'll be no worse off for the remainder of the night breathing just like you would if you didn't wear the mask that night. The force required during inhaling would be only slightly more then normal. The force required to exhale actually would be less then a standard CPAP machine since you aren't going against the flow.

Check with your CPAP machine to see if it can run directly on a battery. My Fisher & Paykel CPAP normally runs on 120VAC, but also has a 12VDC port for running on battery, camper/RV, etc. You can get a battery backup like this one (just an example, not necessarily endorsing it) that will power your machine for up to 15 hours.

I wouldn't go the UPS route for 2 reasons. One is that you're adding additional inefficiencies with the AC->DC->AC conversion process. Plus a device intended specifically for CPAP machines will be easier to help get insurance to pay for, where a UPS they will definitely balk at. They will still probably balk initially at the battery backup, but at least you'll have more going for you.


cableties
Premium
join:2005-01-27

reply to IowaCowboy
»www.cpapdiscountstore.com/catalo···709.html

I would seriously check with your CPAP manufacturer and those that use the apnea devices *BEFORE* going the route of a computer battery backup.

The 865Watt is not realistic in 168minutes runtime (it does vary) and less than 3 hours is not a full night sleep. Ofcourse will you awaken when you hear it beep-beep-beep? Or the fan runs (the new APC models have fans that are loud when recharging).

BTW-I had that model and it ran down to 15mins in less than an hour (I only had a 108Watt Xbox and 25Watt LCD connected...it should have lasted 2-3hours...I shutdown much sooner than that)
--
Splat



jjoshua
Premium
join:2001-06-01
Scotch Plains, NJ
kudos:3

reply to IowaCowboy
For certified medical devices, you should check with the device manufacturer only. Don't rely on any other advice.



DarkLogix
Texan and Proud
Premium
join:2008-10-23
Baytown, TX
kudos:3

reply to cableties

said by cableties:

BTW-I had that model and it ran down to 15mins in less than an hour (I only had a 108Watt Xbox and 25Watt LCD connected...it should have lasted 2-3hours...I shutdown much sooner than that)

Ya the backups units don't get as near the extremes as the smartups do

even then I bet that 168 number is just calculated and it'll likely never actually last that long.


fifty nine

join:2002-09-25
Sussex, NJ
kudos:2

reply to IowaCowboy
If the CPAP machine suffocates you when power goes out, I don't see how the FDA or other Gov't agency would allow its use and sale.

I am pretty sure that if/when power goes out it will simply fail "open" and you will still be breathing but sans pressure.

However if you really want a UPS I would recommend a pure sine wave on-line UPS such as one of the small liebert UPSes from refurbups.com

I would not use the humidifier with the UPS.



alkizmo

join:2007-06-25
Pierrefonds, QC
kudos:1

reply to IowaCowboy
As DarkLogix has said in his point #2, a UPS won't last long unless it's designed to last long. UPS's are designed to last a little while but with a lot of power.

Example:

My UPS can supply 810W
It can sustain 9 minutes at half load.

810w * 50% * 9 minutes = 3645 Watt*Minute

Your thing takes 90 watts.
3645 / 90 = 40.5 minutes

That's way too short.



cdru
Go Colts
Premium,MVM
join:2003-05-14
Fort Wayne, IN
kudos:7

reply to cableties

Ofcourse will you awaken when you hear it beep-beep-beep? Or the fan runs (the new APC models have fans that are loud when recharging).

I'm going to bet that a UPS fan is going to be an issue when a CPAP machine is a fan in and of itself.

said by cableties:

The 865Watt is not realistic in 168minutes runtime (it does vary) and less than 3 hours is not a full night sleep.
...
BTW-I had that model and it ran down to 15mins in less than an hour (I only had a 108Watt Xbox and 25Watt LCD connected...it should have lasted 2-3hours...I shutdown much sooner than that)

Many CPAP machines run well below 1A at 120v. Newer models may be as low as 30 watts after startup. Even if you get 1/2 the battery capacity for usable runtime, you better get at least 168 minutes out of it, if not significantly longer.


workablob

join:2004-06-09
Houston, TX
kudos:1

reply to IowaCowboy
I have sleep apnea too and never worried about this as if air got thin I would wake up.

Are you saying if you stop breathing you will not wake up gasping for air and just quietly suffocate?

Dave



workablob

join:2004-06-09
Houston, TX
kudos:1
Reviews:
·Comcast

reply to cdru

said by cdru:

said by IowaCowboy:

I am thinking of purchasing a UPS back-up power supply to plug my CPAP machine (medical device for sleep apnea) into as I use a full face mask and worry about an interruption in the power supply causing me to suffocate.
...
I have severe sleep apnea to the point of being life-threatening.

Have you ever heard of someone dying from suffocation due to a power outage while on a CPAP machine? They are designed for such instances. Worst case, if the power goes out, you'll be no worse off for the remainder of the night breathing just like you would if you didn't wear the mask that night. The force required during inhaling would be only slightly more then normal. The force required to exhale actually would be less then a standard CPAP machine since you aren't going against the flow.

Check with your CPAP machine to see if it can run directly on a battery. My Fisher & Paykel CPAP normally runs on 120VAC, but also has a 12VDC port for running on battery, camper/RV, etc. You can get a battery backup like this one (just an example, not necessarily endorsing it) that will power your machine for up to 15 hours.

I wouldn't go the UPS route for 2 reasons. One is that you're adding additional inefficiencies with the AC->DC->AC conversion process. Plus a device intended specifically for CPAP machines will be easier to help get insurance to pay for, where a UPS they will definitely balk at. They will still probably balk initially at the battery backup, but at least you'll have more going for you.

This.


bobrk
You kids get offa my lawn
Premium
join:2000-02-02
San Jose, CA
Reviews:
·SONIC.NET

said by workablob:

said by cdru:

said by IowaCowboy:

I am thinking of purchasing a UPS back-up power supply to plug my CPAP machine (medical device for sleep apnea) into as I use a full face mask and worry about an interruption in the power supply causing me to suffocate.
...
I have severe sleep apnea to the point of being life-threatening.

Have you ever heard of someone dying from suffocation due to a power outage while on a CPAP machine? They are designed for such instances. Worst case, if the power goes out, you'll be no worse off for the remainder of the night breathing just like you would if you didn't wear the mask that night. The force required during inhaling would be only slightly more then normal. The force required to exhale actually would be less then a standard CPAP machine since you aren't going against the flow.

Check with your CPAP machine to see if it can run directly on a battery. My Fisher & Paykel CPAP normally runs on 120VAC, but also has a 12VDC port for running on battery, camper/RV, etc. You can get a battery backup like this one (just an example, not necessarily endorsing it) that will power your machine for up to 15 hours.

I wouldn't go the UPS route for 2 reasons. One is that you're adding additional inefficiencies with the AC->DC->AC conversion process. Plus a device intended specifically for CPAP machines will be easier to help get insurance to pay for, where a UPS they will definitely balk at. They will still probably balk initially at the battery backup, but at least you'll have more going for you.

This.

Concur. I can breath fine with mine turned off. Plus whenever it gets dislodged or leaks, I wake up.


workablob

join:2004-06-09
Houston, TX
kudos:1
Reviews:
·Comcast

said by bobrk:

Concur. I can breath fine with mine turned off. Plus whenever it gets dislodged or leaks, I wake up.

Same here. I think the OP is worrying for nothing.

OP, Talk to your doctor just to be sure.

Dave


bobrk
You kids get offa my lawn
Premium
join:2000-02-02
San Jose, CA

Definitely, there are a lot of different masks out there but I think they all work the same way.



tmh

@comcastbusiness.net

reply to IowaCowboy
Seriously, it looks like we're overthinking this problem.

Why not buy a power failure alarm, that's loud enough to wake you up when the juice goes out?

You can buy one for under 20 bucks (»www.acehardware.com/product/inde···55761191).



cdru
Go Colts
Premium,MVM
join:2003-05-14
Fort Wayne, IN
kudos:7

reply to workablob

said by workablob:

Are you saying if you stop breathing you will not wake up gasping for air and just quietly suffocate?

Since you have apnea, you should know the answer whether you wake up or not when you quietly start to suffocate.


cdru
Go Colts
Premium,MVM
join:2003-05-14
Fort Wayne, IN
kudos:7

reply to fifty nine

said by fifty nine:

I would not use the humidifier with the UPS.

That likely is not an option that can be switched depending on power source. Either run it 100% of the time, or don't run it at all.